First test on the new CO2 laser
Page 1 of 1
First test on the new CO2 laser
On friday 12th march, I installed the new CO2 laser from Access Laser Co. It is a Merit-S, pre-stabilized, temperature tuned CO2 laser. Both the laser and the RF driver are water cooled. I connected the laser body temperature sensor to the auxiliary input of the chiller, to directly stabilize the laser temperature.
After switching on the laser, it takes approximately half an hour to stabilize. Maximum power is 9.4W.
I placed a 3" focal lenght lens approximately 40cm from the laser output and the Primes beam profiler at 45cm from the lens. This is necessary to increase the size of the beam, since the beam profiler cannot measure beams smaller than 10mm in diameter.
First figure shows the measured beam intensity distribution: it is fairly good. Second and third images show the profiles in the X and Y directions respectively together with Gaussian fits. The beam radius measured by the profiler agrees very well with the results from the fit: Primes (wx=7.33mm and wy=7.66mm); Gaussian fit (wx=7.16mm and wy=7.60mm). As a final crosscheck, I propagated the beam from the laser to the profiler using ABCD matrices and Gaussian optics: for an input beam of w0=(1.2 - 1.3)mm I get w1=(7.2 - 7.5)mm.
After switching on the laser, it takes approximately half an hour to stabilize. Maximum power is 9.4W.
I placed a 3" focal lenght lens approximately 40cm from the laser output and the Primes beam profiler at 45cm from the lens. This is necessary to increase the size of the beam, since the beam profiler cannot measure beams smaller than 10mm in diameter.
First figure shows the measured beam intensity distribution: it is fairly good. Second and third images show the profiles in the X and Y directions respectively together with Gaussian fits. The beam radius measured by the profiler agrees very well with the results from the fit: Primes (wx=7.33mm and wy=7.66mm); Gaussian fit (wx=7.16mm and wy=7.60mm). As a final crosscheck, I propagated the beam from the laser to the profiler using ABCD matrices and Gaussian optics: for an input beam of w0=(1.2 - 1.3)mm I get w1=(7.2 - 7.5)mm.
Alessio- Posts : 127
Join date : 2008-11-07
Age : 52
Location : Rome
Similar topics
» Chiller and Laser
» Pre-amp Frascati in the laser lab
» TCS Laser protection
» WI CO2 laser noise
» New Gaussian Fit on the temperature curve due to the YAG laser
» Pre-amp Frascati in the laser lab
» TCS Laser protection
» WI CO2 laser noise
» New Gaussian Fit on the temperature curve due to the YAG laser
Page 1 of 1
Permissions in this forum:
You cannot reply to topics in this forum
|
|