Yesterday, with Mauro I., we tried to increase the gain of the Frascati pre-amp. We simply changed the feedback resistor, as shown in the figure.
At first, we used a 2.5kOhm resistor, but it turned out that the pre-amp is not working anymore: no output signal at all. Probably this is due to the fact that increasing the feedback resistor too much unbalances some signal levels.
After several tests, it turned out that it is possible to increase the gain only by a factor of 2.
So now the gain is approximately 200 (measured with an oscilloscope, to be better measured in TOV), the noise at the output is around 200nV/Hz1/2, which means that the input referred noise is around 1nV/Hz1/2. All this must be checked in TOV.
At first, we used a 2.5kOhm resistor, but it turned out that the pre-amp is not working anymore: no output signal at all. Probably this is due to the fact that increasing the feedback resistor too much unbalances some signal levels.
After several tests, it turned out that it is possible to increase the gain only by a factor of 2.
So now the gain is approximately 200 (measured with an oscilloscope, to be better measured in TOV), the noise at the output is around 200nV/Hz1/2, which means that the input referred noise is around 1nV/Hz1/2. All this must be checked in TOV.
Alessio- Posts : 127
Join date : 2008-11-07
Age : 52
Location : Rome
Re: Frascati pre-amp: gain increased to 200
I think that we should compare the noise with gain of 100 with the noise with gain of 200.
Why do you want increase the gain ?
Why do you want increase the gain ?
mdipaolo- Posts : 106
Join date : 2008-11-05
Similar topics
» Pre-amp Frascati in the laser lab
» PD 12595 & Low noise pre-amp Frascati
» Low Noise Pre-amplifier (Frascati)
» PD 12595 & Low noise pre-amp Frascati
» Low Noise Pre-amplifier (Frascati)
Page 1 of 1
Permissions in this forum:
You cannot reply to topics in this forum