Converter voltage/current - Simulation practice for DT470 - SD
Converter V/I: (Basic scheme)
RL: Static impedance DT470 - SD (RL replacing the probe).
The values of the resistors are such that Vin is the order of about 1V, and IL is equal to 10*10-6A.
This value is necessary for the operation of DT470 - SD.
Stability control in the current
Indirect control:
Measure VL: 1.00V
Measure RL: 100kOhm
Calculated value (IL): 10.0*10-6A
The stability in time of the current (IL), was confirmed.
Direct control:
Direct measurement of the current in real time for 10min.
Stable value over time: 10.0*10-6A
The value of Vin was calibrated in both cases.
Conclusions:
- Positive results in this simulation practice
Next Steps:
- Direct test with the probe DT470 - SD
RL: Static impedance DT470 - SD (RL replacing the probe).
The values of the resistors are such that Vin is the order of about 1V, and IL is equal to 10*10-6A.
This value is necessary for the operation of DT470 - SD.
Stability control in the current
Indirect control:
Measure VL: 1.00V
Measure RL: 100kOhm
Calculated value (IL): 10.0*10-6A
The stability in time of the current (IL), was confirmed.
Direct control:
Direct measurement of the current in real time for 10min.
Stable value over time: 10.0*10-6A
The value of Vin was calibrated in both cases.
Conclusions:
- Positive results in this simulation practice
Next Steps:
- Direct test with the probe DT470 - SD
Alessandro- Posts : 42
Join date : 2008-12-11
Location : Roma
Re: Converter voltage/current - Simulation practice for DT470 - SD
For avoid auto-oscillation of positive fedback you can try to put in parallel to R4 one capacitor.
mdipaolo- Posts : 106
Join date : 2008-11-05
Similar topics
» Converter Voltage/Current - Circuit assembly
» Information from Datasheet Thermometer DT470 SD
» Data acquisition (DT470 SD - Calibration Curve) - Channels DF0-DF1-DF2-DF3
» Test real-time data acquisition with DT470-SD
» TCS Radiative Cooling
» Information from Datasheet Thermometer DT470 SD
» Data acquisition (DT470 SD - Calibration Curve) - Channels DF0-DF1-DF2-DF3
» Test real-time data acquisition with DT470-SD
» TCS Radiative Cooling
Page 1 of 1
Permissions in this forum:
You cannot reply to topics in this forum
|
|