(Viviana, Alessio)

This morning we worked on the NI intensity stabilization. We have
set the laser temperature to about 20.9 deg, that gives a
good stability of the output power.
Figure 1 shows a comparison between the open loop noise spectrum
(pink) and closed loop spectrum (black). The reduction factor is the
same as the WI bench (our reference).
We also investigated the origin of the high frequency bumps (from
100Hz on). We compared the closed loop spectrum with the dark noise of
the photodiodes:
1) figure 2: against the out-of-loop PD, it does not seem to be the cause
2) figure 3: against the in-loop PD, it seems to be the cause
3) figure 4: to understand if the culprit is the PD or the low noise
amplifier, we connected the amplifier to the out-of-loop PD. thus it
seems that the source is the pre-amplifier.
We want to underline that, since these bumps are at high frequency, they are not important for the sensitivity.
NI intensity stabilization 26343_10
NI intensity stabilization 26343_11

NI intensity stabilization 26343_14

NI intensity stabilization 26343_13
Alessio