WI TCS beam alignment on ITM
(Viviana, Alessio, Laura, F.Frasconi)
On tuesday afternoon we performed some operations required to allow closing of the WI tower.
1) placed a lens in front of the cross-hair to increase intensity of the arms of the cross.
2) aligned CO2 beam onto the cross-hair.
3) replaced hot mirror with 95%R beam splitter (to allow positioning of the IR camera, not yet done).
On wednesday morning, with Franco, we re-installed mirror B. Finally, we checked the alignment of the CO2 beam on mirrors A and B by looking at the position of the cross-hair laser on the mirrors. We took a picture of the CO2 ring on a screen placed on mirror B with the handy thermal camera. In this way we could check if there is any clipping from mirror A and the centering on mirror B. We found some clipping on the left side on mirror B, thus we reduced a bit the dimensions of the annulus by moving L5.
Here the distances between some optical component on the NI and WI:
NI--> L5-BS = 17.5cm; BS-L4 = 5cm; L4-AXI = 4.8cm
WI--> L5-BS = 17.5cm; BS-L4 = 4cm; L4-AXI = 5cm.
We re-checked the alignment in tower and the image of the CO2 ring on mirror B (see figure 1).
We aligned the cross-hair on the reference cross drawn on the screen mounted on the reaction mass. We left the motors at half of their dynamics.
On tuesday afternoon we performed some operations required to allow closing of the WI tower.
1) placed a lens in front of the cross-hair to increase intensity of the arms of the cross.
2) aligned CO2 beam onto the cross-hair.
3) replaced hot mirror with 95%R beam splitter (to allow positioning of the IR camera, not yet done).
On wednesday morning, with Franco, we re-installed mirror B. Finally, we checked the alignment of the CO2 beam on mirrors A and B by looking at the position of the cross-hair laser on the mirrors. We took a picture of the CO2 ring on a screen placed on mirror B with the handy thermal camera. In this way we could check if there is any clipping from mirror A and the centering on mirror B. We found some clipping on the left side on mirror B, thus we reduced a bit the dimensions of the annulus by moving L5.
Here the distances between some optical component on the NI and WI:
NI--> L5-BS = 17.5cm; BS-L4 = 5cm; L4-AXI = 4.8cm
WI--> L5-BS = 17.5cm; BS-L4 = 4cm; L4-AXI = 5cm.
We re-checked the alignment in tower and the image of the CO2 ring on mirror B (see figure 1).
We aligned the cross-hair on the reference cross drawn on the screen mounted on the reaction mass. We left the motors at half of their dynamics.
laura- Posts : 8
Join date : 2009-11-05
Similar topics
» NI TCS beam alignment on ITM
» NI ring alignment shift 28 Jan 2009
» Phase Camera 1: Ring Alignment Shift 19 Jan 09
» Thermal camera installation on NI TCS bench and alignment checks
» CH beam splitters replacement
» NI ring alignment shift 28 Jan 2009
» Phase Camera 1: Ring Alignment Shift 19 Jan 09
» Thermal camera installation on NI TCS bench and alignment checks
» CH beam splitters replacement
Page 1 of 1
Permissions in this forum:
You cannot reply to topics in this forum
|
|